Today I noticed when taking a image the Cooler switches on even if it’s Off and I don’t want cooling right now.
The main problem is that camera reports -5.4C but the real temperature is +9.0C, so the cooler tries to reach
a higher temperature and that’s really difficult..
Anyone else noticed something like this ??
This happens using flatpak KStars Version 3.8.5 Stable and Camera SX-814
Ekos
Cooler off
Camera shows Temp: -8.1C Scope Temp: +7.0C
Press “Preview” and Cooler goes On
Press “Abort” and Cooler stays On, have to switch off Cooler (in Ekos) otherwise the Cooler stays On.
Press “Looping” and Cooler goes On and Progress stays at “Set Temp to 0C…” until I press “Abort”
If I take a exposure using the Sequence Queue the result is the same.
So no way to take a image if I don’t cool down the Camera to -8.1C.
The funny thing I just discovered is that if I start the exposure from INDI the Cooler stays Off as normal.
Change the temperature value to “–” by scrolling all the way down.
Use KStars Flatpak beta 3.8.6 as it removed the need for wait for temperature set point to reach target value while either in Preview or Looping modes.